Hi all. Following on from my thread asking for opinions on my membrane, I've fitted a bypass with a stopcock so I can divert the pure side into the waste while I'm flushing and a flow meter on both the pure and waste lines.
This means I can monitor the performance of the membrane very accurately.
So:
I was running at 2.5:1 ratio (waste to pure). The pressure at the membrane was 60psi. Incoming water is around 300ppm (varies slightly) water out of the membrane (before DI) was 24ppm.
Supposedly if the pressure is higher, you should get lower tds out. If you reduce the volume going to waste by closing the flush valve slightly this will increase the pressure at the membrane (less water to waste=more water through the membrane=higher pressure). I reduced to a ratio of 1.5:1. The pressure increased to 90psi but the tds out of the membrane almost doubled to 48ppm
This was the exact opposite of what I expected - I thought that increasing pressure would reduce the tds but probably shorten the life of the membrane.
I increased the waste gradually and the tds came down as the pressure reduced and the waste increased. By trial and error I found that the best tds I could get was the 24 I had been getting originally, but at about 65psi and a ratio of slightly less than 2.5:1. Reducing the pressure (and increasing the waste) beyond that made no difference so I am assuming I have found the optimum performance setting for my membrane.
Working out the percentages, I am getting 92% rejection which still means the resin has a lot of work to do to get the 24tds down to 001 (000 seems unattainable)
I'm going to buy a new membrane, probably the HF5 but I'll wait till I can talk to June at Gapswater.
My old RO-man 300gpd used to produce water at 3-4ppm before DI so the performance of my current 40-40 leaves a lot to be desired.
I'm changing resin now about every 8 days (when the tds reaches 009) so I'm hoping I will get less than 009 from a new membrane and then I won't need to use DI, or if I do, it should last a good deal longer.