I just wanted to check with others to see if you agree.. I have just fitted 3 new 100 gpd membranes, filters,and new resin after running the system for nearly 24 hours and prior flushing after taking seperate readings 2 of the membranes were bringing the tds down to 20ppm before the di and 1 membrane could only manage it down to 200ppm would you agree that this membrane is faulty..
thanks
paul
-
It is more likely that it has not been seated properly in it's housing. It would be worth taking it out putting a very little silicone grease on the the O rings and fitting it back in, being careful to push it fully into position.
If this does not improve it then it would have to be a faulty membrane. This is very rare, but can happen occasionally.
-
thank you alex i did take it out and replace it a couple of times but was still getting the same results..then put an old one in from the last lot as they were only a few months old and it was ok the only reason they have been replaced was because the housings all cracked with the recent freeze..
paul
-
Must be the membrane then, unless it is some very obtuse yet to be discovered problem!